Helmets for CPAP
Giacomo Bellani, MD, PhD University of Milan-Bicocca
A.O. San Gerardo Monza (Italy)
Conflicts of Interest
INSTITUTIONAL:
Research grants from:
– Draeger
– Maquet
– Airway Medix
– Chiesi Farmaceutica, SPA
Patent transferal to: Draeger
PERSONAL
Consultancy fee from: Draeger
PEEPvalve
Gas Inlet
Helmets for CPAP
Giacomo Bellani, MD, PhD University of Milan-Bicocca
A.O. San Gerardo Monza (Italy)
CPAP
JAMA 2005 (Dec); 294: 3124-3130
Acute Cardiogenic PE
F = 90 l/min F = 90 l/min
Paw
Pat flow
Airway flow
F = 40 l/min F = 90 l/min
Paw
Pat flow
Airway flow
F = 50 l/min
F = 90 l/min
Paw
Pat flow
Airway flow
F = 120 l/min
F = 210 l/min F = 120 l/min
F = 90 l/min
Expiratory valve resistance
5
7
9
11
13
15
17
19
30 40 50 60 70 80 90 100 110 120
P in
sid
e h
elm
et (
cmH
2O
)
Gas Flow (L/min)
10
Isgrò S et al., Minerva anestesiol 2003
Keep airway pressure stable
High fresh gas flow rates
Low FiO2 (venturi)
High O2
consumption
Expiratory workload
Humidification
Compliance
Then, what do we need?
Time
Pressure
Ppl
Paw
WOB imposed by circuit
What does reservoir do?
Helmet volume (L) 13 22 17
Helmet
compliance
(ml/cmH2O)
PEEP 5 726 1131 826
PEEP 10 456 666 496
PEEP 15 354 503 385
Helmet Hose
Resistance
(cmH2O/Lxsec -1)
0.48 0.6 0.24
PN500 - Harol CaStar - StarMed 4Vent - Rüsh
Figure 2: effect of different CPAP levels on FRC during CPAPH (left side) and CPAPM (right side) in each subject. Each point
represent mean ± sd of FRC between different gas flow rates. Dotted lines refer to subjects with evidence of expiratory muscle
contraction.
* effect of CPAP levels during CPAPH p<0.05; # effect of CPAP levels during CPAPM p<0.05.
0
500
1000
1500
2000
5 10 15 5 10 15
Helmet Mask
F
RC
(m
l)
* #
CPAP levels (cmH2O)
FRC changes
Patroniti N., Foti G., Pesenti A. et al. ICM (2003). 29: 1680-87
Bias Flow in the Head Tent MUST be > 30 L/min.
0
2
4
6
8
10
12
14
20 30 40 50
Bias Flow (L/min)
PC
O2
ins
p (
mm
Hg
)
0
200
400
600
800
1000
1200
TV
(m
l)
PCO2insp.
TV
Head Helmet versus Face Mask for Non Invasive CPAP : a physiological study
Patroniti N., Foti G., Manfio A., Coppo A., Bellani G., Pesenti A.
ICM September 2003 29:1680-1687
0
2
4
6
8
10
12
14
16
0
5
10
15
20
25
30
35
40
45
50
Face Mask 60 L/min 10 L/min 30 L/min Ventilator
PetCO2
PiCO2 (m
mH
g)
VE (
L/m
in )
Taccone P. et al. Crit Care Med (2004)
Conclusion
Helmet CPAP is an effective way to provide noninvasive CPAP
As effective as face mask (less O2 consuption, higher FiO2)
Some data suggest a better comfort than face mask
Rebreathing has to be taken into account (avoid the use of a MV to provide CPAP)
Thank you!